Re: Use of imager spacing tag (0018,1164) and/or pixel spacing tag (0028,0030) for measurement /cali
This topic and its replies were posted before the current version of LEADTOOLS was released and may no longer be applicable.
#1
Posted
:
Monday, September 28, 2009 11:44:42 PM(UTC)
Groups: Registered
Posts: 22
We use Leadtools v15 to perform measurements (distance) on images. The modality provides calibration information in the above two tags (or one of these).
As i understood from leadtools, tag 0028,0030 is used for calibration of the ruler.
However i have found a number of issues on that make it a bit confusing.
Attached i have a test image for pixel spacing. When i use this image in the Leadtools medical viewer, the calibration is not correct. (according the image).
My question exact: Which if these tags should be used?
#2
Posted
:
Tuesday, September 29, 2009 6:25:25 AM(UTC)
Groups: Guests
Posts: 3,022
Was thanked: 2 time(s) in 2 post(s)
There's a clarification document on the NEMA web site which discusses this. You can find it here:
ftp://medical.nema.org/medical/dicom/final/cp586_ft.pdf
In any case, there's no attached file, if you want to attach it again, please put it in a ZIP or a RAR file and do NOT use the preview feature.
LEADTOOLS Support
Medical
Medical SDK Questions
Re: Use of imager spacing tag (0018,1164) and/or pixel spacing tag (0028,0030) for measurement /cali
You cannot post new topics in this forum.
You cannot reply to topics in this forum.
You cannot delete your posts in this forum.
You cannot edit your posts in this forum.
You cannot create polls in this forum.
You cannot vote in polls in this forum.